如题,小弟用的ITO的型号是SPI USA 15-30 ohm;
铁氰化钾溶液为1m M的K3Fe(CN)6 in 0.1 M的KCl尝试的清洗方法有(但始终CV图没有改善):(1)CH2Cl2 sonication -> methanol sonication 10 min -> 0.5 M K2CO3 in methanol:H2O (V/V 3:1) 45 min sonication-> rinsed with copious H2O-> N2 /140 ℃ dry it (Langmuir,2011,27,2545-2552)(2)acetone sonication 10 min-> isopropyl alcohol sonication 10 min -> RCA mixture (1:1:5 NH4OH:H2O2:H2O) sonication 10 min -> boiling in 80 ℃ H2O 30 min-> rinsed with copious H2O?--》N2 dry it (J.Phys. Chem. B 2001, 105, 4270- 4276)CV条件为:init E(mV)=0? ?? ?? ?? ?? ???High E(mv)=200? ?? ?? ?? ?? ?? ?Low E(mV)=-800? ?? ?? ?? ?? ???Init P/N=P (positive)? ?? ?? ?? ?? ? N.O. of segments=6? ?? ?? ?? ?? ? sample interval (mV)=1? ?? ?? ?? ?? ?? ?Quiet time (sec)=2? ?? ?? ?? ?? ? sensitivity(A/v)=10 uA/V得到的结果如图:尝试了10次了,难道就是表面没清洁干净,干燥后滴水都能铺开,表面难道还不够干净??所有的溶液也重新配了,实在找不到原因了,文献结果重现不出来,困惑ing,请各位高手指点迷津,不胜感激!!~cleaned bare ITO.jpg